At the moment I am repairing a Sennheiser MKH40 microphone.
There is something not clear in the calibration procedure.
The Sennheiser service manual says:
"Check the voltage between the measuring points A(+) and B(-).
Desired voltage is 0 V.
If the value is not 0 V, rotate the coil core of L1 until the voltage is 0V."
But... in the schematic point A is +36.5 V and point B is +37.5 V, so not exactly what you would call 0 V...
I would expect the voltage on point A to be +37,5 V also, because at the top of D3 the voltage is +42 V, and at the bottom of D4 the voltage is +33V. So the difference is 9 V and IMHO this could mean that the voltage between D3 and D4 could be 33 + (1/2 * 9) = 37.5 V.
Am I missing something, or are the voltages mentioned in the service manual (+36.5 and +37.5 V) correct?
Could it be that the differential voltage would give a different result than the voltage on these points referred to ground?
There is something not clear in the calibration procedure.
The Sennheiser service manual says:
"Check the voltage between the measuring points A(+) and B(-).
Desired voltage is 0 V.
If the value is not 0 V, rotate the coil core of L1 until the voltage is 0V."
But... in the schematic point A is +36.5 V and point B is +37.5 V, so not exactly what you would call 0 V...
I would expect the voltage on point A to be +37,5 V also, because at the top of D3 the voltage is +42 V, and at the bottom of D4 the voltage is +33V. So the difference is 9 V and IMHO this could mean that the voltage between D3 and D4 could be 33 + (1/2 * 9) = 37.5 V.
Am I missing something, or are the voltages mentioned in the service manual (+36.5 and +37.5 V) correct?
Could it be that the differential voltage would give a different result than the voltage on these points referred to ground?