In that case 1/10th of 2MHz would be around 15 meters for the cable. So, why the service manual says to terminate with 50 Ohms for calibration?
1) You're calibrating it, so rules of thumb (educated short cuts) do not apply.
2) What's the highest frequency of a 2 MHz square wave?
 
If the SKU has 50 ohm source impedance and expects a 50 ohm load termination that will cause a 6 dB drop in level. Calibrating without the proper load/termination will not deliver correct level.

JR
 
In that case 1/10th of 2MHz would be around 15 meters for the cable. So, why the service manual says to terminate with 50 Ohms for calibration?
Because that becomes part of the output circuit and affects the output voltage; their specified O/P impedance is apparently 50Ω.
 
In that case 1/10th of 2MHz would be around 15 meters for the cable. So, why the service manual says to terminate with 50 Ohms for calibration?
Because the level depends on the load impedance. For equal source and load Z, the signal is attenuated 6dB.
It's important for level calibration, although the frequency response would be correct even with a large mismatch.
 
The usual RoT is when the cable length exceeds 1/10th the wavelength of the signal.
In that case 1/10th of 2MHz would be around 15 meters for the cable. So, why the service manual says to terminate with 50 Ohms for calibration?
Unterminated connection is OK (insignificant SWR) for a short cable at 2 MHz but puts a different level out. Calibration is about precision.
 