Major changes that someone did were: OA2 changed to NE5532 OA3 changed to LME49720NA OA4 changed to NE5532, most of the Rs in the audio path were changed to metal film on the main board but not on the VCA cards, caps in the audio path were changed to film style.
R45 gives voltage trim of +/-10V to +/-12.3V
Test setup APSYS2 -5dBu Unbal - GND output.
Front panel controls: switch in the out position threshold just changed from yellow to red, Compression 1.0 See first scope capture.
If I remove the right channels VCA the signal cleans up. Second scope capture.
Bench supply actually seems to be causing a HF oscillation 80kHz
Also weird one both VCA boards are in, driving left channel and put scope probe on R32 where it connects with P3 pin 3 signal cleans up.