If it asks for 1 kHz at 0 dB, does this mean a 1 volt peak to peak on my scope hooked up to the signal generator? And +1 dB = ? volts p-p?
Okay, in this case, 0db = .775Vrms. That is your reference voltage.
.775 X 1.414 X 2 = peak to peak volts on the scope.
To find what voltage corresponds to +1db, use the formula:
db = 20[log(V1/V2)]
I'd just start plugging in values for V1, using .775V for V2, until you get the correct db result (+1). Or you could rearrange the above equation to solve exclusively for V1. :roll:
Don't forget to convert to P-P if you're using the scope to measure levels.
The Sifam VU or equivalent reads 1.23Vrms at 0db rather than .775Vrms because it is referenced to +4db, which is what you'll get if you plug 1.23V and .775V into the above equation.
To recap, .775Vrms is the 0db reference for all calibration procedures on the 1176.
Dean