I finally get my new rev J- twins alive and needed to calibrate those. It was really hard to get easy or good calibration instructions to this rev J and so I made new kind calibration method. Much easier for me... need digital multimeter and output should connect to for example to mixer or some kind of that (to balanced input anyway).
First, you need to know... I have cheap DC- "VU- meter". Needed diode bridge and extra 1 kohm resistor to parellel with R51 (meter board) to show 0VU right.
1. Okay, run the unit for 10 minutes to stabilize components' temperature.
2. Make Q-Bias adj. Set ratio to 1:20, meter mode to OFF. Set release and attack fully CW, IN POT CCW and OUT POT 3/4 CW.
Then apply a 0dBu (0.775 Vrms) 1kHz sine wave to the input. Adjust IN POT get +1VU (0.869 Vrms) to the outputs digital multimeter. Start with the Q-Bias trimmer fully CCW. Then turn the trimmer CW until the digital multimeter reads 0VU (0.775 Vrms). Now the FET is sligtly in conduction.
3. GR Meter Tracking and ZERO. Set ratio to 1:20. Set meter mode to GR. Set input middle and output to CCW. Set release to fully CW and attack CCW.
A - With no signal applied, adjust VR55 (ZERO adj) so that the VU- meter reads 0VU.
B - Apply a 0dBu (0.775Vrms) 1kHz sine wave signal to the input.
C - Switch the meter mode to OFF (comp. bypass).
D - Turn the output level control until the outputs Digital MultiMeter, DMM reads 0dBu (0.775Vrms).
E - Switch the meter mode to GR.
F - Turn the input level control until the DMM reads -10dBu (0.245Vrms).
G - Repeat steps C to F until DMM show (0.775Vrms) on bypass- mode and (0.245Vrms) on GR- mode.
H - Leave the meter mode to GR.
I - Adjust VR54 (TRACKING) so that VU- meter reads -10VU.
J - Turn off signal.
K - Adjust VR55 (ZERO adj) so that the VU- meter reads 0VU.
L – Turn signal on.
M- Repeat steps I to L until VU- meter show right in both cases.
That’s it… so simple, and compressors work as old trains toilet.
BTW. What you think about this kind calibration steps?